Author Topic: SIREMOBIL 2000  (Read 101 times)
JGMORILLO
Newbie
*
Posts: 1


View Profile
« on: July 29, 2010, 11:50:37 AM »

AFTER ELONGED USE, ERROR 5017 SHOWED AND NO XRAY OR FLUORO AGAIN.
PULSE FLUORO MAKE XRAY INDICATOR TURN ON FOR A MOMENT AND OFF AGAIN
AFER CHECK ALL FUSES AND SUPPLIES WITHOUT FIND ANY BAD, ERROR MOVE TO 001E AND NOW NO INDICATOR OF XRAY AT ALL

ANY HELP WILL BE HIGHLY APPRECIATED
Logged
vesper
Newbie
*
Posts: 31



View Profile
« Reply #1 on: July 29, 2010, 11:53:44 PM »

Error 5017
----------
Meaning: Minimum kV not attained
Cause: Controller D202 monitors the min. permissible value of the high voltage. If this value is not attained, an error is displayed.
Active: during radiation
Possible trouble / help : No actual kV value present - Inverter not functioning - No drive pulses present.


Error 001E
----------
Meaning: D11 +15V or -15V out of tolerance
Cause: Message from power supply board D11. The +15 V or -15V voltage is out-of-tolerance.
Active: Constantly
Possible trouble / help : Power supply board D11 defective - Connection between D11 and D6 is broken.

Good luck.